alrha said:
i change the sediment filter when i notice decreased flow rates.
i change the carbon filter when i see it clogged (orange).
the DI is supposedly supposed to change color when exhausted, i never saw it change color, but changed it after one year as well as the RO membrane.
my tap water tests at ~45 TDS and my RO/DI tests at ~1 TDS.
A good rule of thumb is to replace your sediment filter and carbon block after six months. A more precise way to maximize the useable life of these two filters is to use a pressure gauge to identify when pressure reaching the membrane starts to decline. This is your indication one or both of the filers is beginning to clog.
Also be cognizant of the chlorine capacity of the carbon block. The Matrikx+1 for example will remove >90% of chlorine from 20,000 gallons of tap water presented at 1 gpm. Original equipment suppliers commonly provide carbon cartridges rated at 2,000 to 6,000 gallons.
Regarding your RO membrane and DI resin, use your TDS meter to measure, record, and track the tds (expressed in parts per million) in three places:
1. Tap water
2. After the RO but before the DI
3. After the DI.
The TDS in your tap water will likely range from about 50 ppm to upwards of 1000 ppm. Common readings are 100 to 400. So for sake of discussion, let's say your tap water reads 400. That means that for every million parts of water, you have 400 parts of dissolved solids.
How do we go about getting that TDS reading down to somewhere near zero?
If you do some experimenting with your TDS meter, you'll note that your sediment filter and carbon block filter do very little to remove dissolved solids. So with your tap water at 400 ppm, you can measure the water at the "in" port on your RO housing and you'll see its still approximately 400 ppm.
The RO membrane is really the workhorse of the system. It removes most of the TDS, some membranes to a greater extent than others. For instance, 100 gpd Filmtec membranes have a rejection rate of 90% (i.e., they reject 90% of the dissolved solids in feed water). So the purified water coming from your 100 gpd membrane would be about 40 ppm (a 90% reduction). Filmtec 75 gpd (and below) membranes produce less permeate, but have a higher rejection rate (98%).
If you measure the TDS in your system after the RO membrane, and before the DI housing, you'll be able to measure the rejection rate of your membrane. A declining rejection rate is a good way to tell when your membrane needs to be replaced.
After the RO membrane, water will flow to your DI housing. DI resin in good condition will reduce the 40 ppm water down to 0 or 1 ppm. When the DI output starts creeping up from 0 or 1 ppm to 3 ppm, 5 ppm, and higher, you know that your resin needs to be replaced.
Sometimes people complain that their DI resin didn't last very long. Usually the culprit is a malfunctioning RO membrane sending the DI resin "dirty" water. This will exhaust the resin quicker then would otherwise have been the case.